Issue 34 is a good one, and you will see me refer to that in issue 47 during a discussion about measuring EGTs for very short dyno runs. Be careful not to interpret the rise times in issue 34 literally; yes the fastest was on the order of 7 seconds, if memory serves, but that was to 1600 degrees I think. Anyway, the important thing is that some are faster than others. For my work, the importance was that the SPA thermocouple responded in half the time as most of the others. It was like the SPA was head and shoulders above the rest in speed, then the bulk of the market followed at about double the rise time, then at the very end was a couple of devices, one especially, that were very very slow. One didn't even acheive the target temperature. you'd be surprised who the slow ones were.
keep in mind that this is pretty old data, so the themocouples tested in the article may have been improved, I don't know.
There are four major sources of error in a type K pyrometer:
1. gauge speed. the gauge itself has to respond to the signal generated by the thermocouple. Analog gauges can be fast, below or about 1 second, but watch out; some respond slower than others. I think the Westachs are among the faster gauge faces out there.
2. Thermocouple speed. Since the primary K junction has to be protected (usually encased with some ceramic type material), the protection itself has a thermal rise time, mass, etc. Very Heavy protection generally means very slow; you can see this evidence by inspecting the tip itself. As issue 34 points out, some are slow and some are fast.
3. gauge calibration. All type K pyrometers measure the difference in temperature between two locations. you are not measuring the absolute temperature of the exhaust manifold, for example, you are measuring the difference between this point and some other point. how a gauge deals with this second point, and the error it introduces, is important. there are three ways I know of (to deal with this error) that are used typically in gauges available to us:
a) the secondary junction is placed under the hood. The Cheaper thermocouples do this, and is evident by the use of ordinary copper wire under the hood, typically emerging several inches from the main probe. No matter where the second K junction (boundary between "K" wire and copper wire) is, the gauge has to make an assumption about the temperature of this junction. This is because the gauge responds to the difference in temperature between the two locations. whatever assumption this is will be wrong by an amount equal to the temperature variation under the hood. for example, if the gauge assumes 100 degrees under-hood temp, and its a cold day, your EGT might be 50 degrees cooler than the gauge says. or, if its a hot day and you are towing hard, under hood temps might get very hot, and your actual EGTs might be 75 degrees hotter than the gauge reads.
b). the 2nd junction is placed in the cab. This is done by the use of very long type K wire (the unweildy and stiff wire) that extends into the cab. In this case, the temperature of the 2nd junction is much more controlled, and the gauge itself can be calibrated with the assumption that if you are driving the truck, the in-cab temperature is probably 75 degrees or so. So all the gauge has to do is assume that the 2nd junction is at 75 degrees, for example. The gauge reading will be wrong, but only by an amount that the cab temperature differs from 75 degrees. I think Westach uses this method; others probably do as well.
c) electronic temperature compensation. this is overkill for all practical purposes, but it is the ultimate in accuracy. The SPA gauge does this -- inserts circuit that actually measures the temperature of the 2nd junction, much like a Fluke meter would, for example, and calibrates itself against that. This produces extraordinary accuracy within 1 degree or less.
4. install location of the primary junction. what ever location you pick, there are errors. if you go post turbo, then distance from the turbo will matter and there will always be some correlation as to what a safe maximum is. if you go pre-turbo, then the location on the manifold matters, the manifold itself matters (stock, ATS, whatever), etc. there are lots of recommendations out there and many experiments have been done, and still not all agree on the best location for mounting the thermocouple.
well that was a bit long. sorry. the point is to understand the souces of error and know your gauge's accuracy. In general, most of us try to eliminate as much error as possible within our means, and get as close as possible to the physical evidence that represents the vulnerability we are out to mitigate. For most of us, that means, a fast gauge, minimally an in-cab location of the remote K junction, and a fast thermocouple mounted in the hottest part of the exhaust manifold.