Steve, based on some recent experimenting with a two digital guages, copper wires, a terminal block, several K-type thermocouples and thermocouple inter connect wire, I would suggest using thermocouple wire and not the copper between the two instruments.
I used copper wire between meter(1) and the T-block and connected a thermocouple connected on the opposite side of the terminal block. Meter(2) had a K-type thermocouple connected directly to the the meter. Both units read within 1 degree of the ambient temperature. When I measured a known temperature (98. 6) meter(1) read about 106 degrees while meter(2), the control, read 97. 8. When I replaced the copper wire between the meter and the T-block, both meters again read within 1 degree of each other at both known temperature measurement points.
I'm sorry that I don't have enough data determine if this is progressive with temperature (66 to 98 = 8 degree error therefore 66 to 366 = 80 degree error?).
The T-block was to allow switching between two thermocouples using copper wire between the T-block and the switch. Only one side of each thermocouple needs to be switched. The other side (obviously needs to be the same side) is daisy chained to the terminal to the meter.
Copper wire was used only for jumpering and runs to the switch.
[This message has been edited by Rattlin (edited 04-07-2001). ]